Agree with above posters: Lab draws can be done on any PICC
4 FR or greater using 10cc syringe. Purpose of PICC is to SAVE VEINS!
From Bard--manufacturer
QUESTION: What syringe size is recommended for use with various catheters?
ANSWER: It is recommended that a 10cc syringe or larger be used to infuse or flush catheters. This includes Pediatric and neonatal catheters. Infusion pressures should never exceed 25 psi because pressures higher than that may damage blood vessels and viscus. A two-pound weight equivalent force on the barrel of a 3cc syringe generates pressure in excess of 25 psi. The same two-pound weight equivalent force on the barrel of a 10cc syringe generates less then 8 psi of pressure. Please refer to the Instructions for Use for further details.
More info here:
http://www.bardaccess.com/faq/q_a.html
Vascular Access Infections in the Outpatient Setting--Power Point Required
www.idlinks.com/slides/ins.ppt
eMedicine - Venous Access Port Problems :
http://www.emedicine.com/aaem/topic472.htm#target1
Patient advocacy alert!!!!
Sounds like many nurses need to check with the manufacturer of IV device being used in their facility for education on devices and use to discuss concerns with prescribing doctors and creating up-to-date policy and procedures!