MarySunshine,
It sounds to me like you are describing the difference between "pole-mount" transducers mounted on an IV pole in a transducer holder with the transducer stop-cock at the level of the phlebostatic axis and "patient-mount"............where the transducer is attached to the patient usually via a garter belt type system or taped either to the patient or to a small towel roll. In
EITHER case...........you are correct in that the transducer stop-cock is supposed to be leveled to the phlebostatic axis when you are recording readings. (We rezero every 4 hours.......or whenever the readings are "iffy".
I haven't noticed a great difference in the patients that I care for between pole-mount and patient mount, but personally I prefer patient mount............as "extra tubing" could tend to increase or amplify catheter whip, spike or overshoot and error.
However amongst the 4 ICU's in my hospital, there is GREAT debate whether pole mount or patient mount is better..........and folks who are convinced that one or the other is ABSOLUTELY the only way to go.
Currently our cardiac unit is the only unit that insists on pole mount, the other 3 all go with patient mounts.
If the transducer is high the reading could be FALSELY low; if the transducer is low, the reading could be FALSELY high........why bother with numbers you can't trust??? I make sure I can trust the numbers by keeping the transducers at the phlebostatic axis and by determining the following the trend between cuff and art line.
We handle arterial lines the same as PA lines in the following expected practices:
See attached PA practice alert.
Meaning the expected practice is:
1. The RN does a square wave test at the beginning of each shift and whenever the art line appears dampened or distorted
2. The transducer stop-cock is level with the phlebostatic axis and the patient is supine with the HOB up no greater than 60 degrees when readings are recorded.
3. I also correlate the arterial line with a manual pressure in both arms at the beginning of my shift; manual pressure in the arm without the art line in it; occlusion pressure in the arm in which the art line is placed. Or manual pressure in both arms if the art line is somewhere other than a brachial or radial artery.
I'll try to look it up in the AACN procedure manual and get back to you tomorrow................

about whether it mentions anything about pole mount versus patient mount, but I doubt it does.
The following member says Thank You: